Topic: Finding replacement for defective oscillator in ADI-2 Rev 1.4
I have recently acquired an old ADI-2 Rev. 1.4 that seems to have a malfunction oscillator. While the device can be operated like described in the manual, all its internal clocks are wildly off and lower than they should be. Maybe somebody with access to a service manual or at least a working device can chime in and help ![]()
I measured some internal signals with my oscilloscope and a 10X probe with ground spring for as little parasitics as possible. With the ADI-2 set as a 48 kHz master, I expect MCLK to be 12.288 MHz or a multiple of that. However, I measure an actual MCLK of about 10.2 MHz, so about 83 % of the nominal value. Fittingly, the SPDIF output and bit clock are about 2.55 MHz, so a quarter of MCLK. The actual values all fluctuate about 5 % as the clock is somewhat unstable.
Syncing to an external 48 kHz SPDIF signal from my Focusrite 18i20 2nd works, and the ADI-2's lock LED is lit. However, none of sample rate indication LEDs light up as the ADI-2 seems to be unable to make sense of the incoming clock. MCLK further reduces to about 6.6 MHz, BCLK and SPDIF output again being 1/4th of that.
I checked all system voltages, and they seem to be OK. Only the "5P" test point gave me a rather low 4.57 V instead of the expected 5 V.
Finally I measured between pins 5 (output) and 4 (GND) of the canned crystal oscillator marked as 25 MHz and got the following:
I got 41.66 MHz instead of the nominal 25 MHz! Also, Vpp seems excessive since the oscillator is supplied with the "5P" voltage. Maybe there is some odd resonance?
My questions are:
- Is my "5P" voltage measuring 4.57 V out of spec? If yes, what may the culprit?
- The crystal oscillator is definitely off, but I do not know if this due to the "5P" voltage supply or just old age. In any case, what are the specs to look for when searching for a replacement?
Best regards,
deadsnare